03-01-2011 06:06 AM
Hello,
I have DAQ Card NI-6036E with connecting block BNC-2120. I would like to measure voltage in multiple channels. I use Labview for programming little aplication. I set the DC voltage on analog output of the same card. When I connect this output voltage to AI01 I see the same values of DC voltage on all reserved AIs even if I don't have rested AIs connected to any source. Why isn't the measeured voltage 0 V on free AIs? If I use another hardware (NI-9215) with the same software application everthing is OK and voltage on free AIs is 0V. Please do you know where could be the problem? Thanks a lot.
03-04-2011 07:03 AM
I detected it was ghosting. http://digital.ni.com/public.nsf/allkb/73CB0FB296814E2286256FFD00028DDF?OpenDocument