Hi Chuck,
There is a bug in DAQmx that causes NI 9219 on-demand thermocouple channels to always use the "high resolution" ADC timing mode when built-in CJC is enabled. It has been reported to R&D as CAR 4BC3NIEV. The workaround is to configure a sample clock, by either calling DAQmx Timing.vi or selecting "N Samples" or "Continuous Samples" for the DAQ Assistant's "Acquisition Mode" setting.
The way you're using the DAQmx Channel property node to set the AI.ADCTimingMode property is almost right but you need to wire the DAQmx Channel to the same DAQmx Task reference that you're passing to DAQmx Start.vi, not a separate DAQmx Task reference. I'll attach a picture (assuming that "T NOx CTR" and such are the names of channels in the "CarDAQ" task):

(By the way, if you right click on the property node input and select
Create»Constant, LabVIEW will put a ring constant on your diagram so that you don't have to look up the numbers.)
Brad
Message Edited by Brad K on 10-31-2007 04:22 PM
---
Brad Keryan
NI R&D