Hi , 


According to the new test requirement, we should generator four signals in the same time. The signals are sine wave with 2.5KHz 50mVpp, 1.1KHz 10mVpp, 1.0KHz 10mVpp, 5.0KHz 100mVpp.  
We also need  to measure the 1.1KHZ + 1.0KHz  +5KHZ measurements in  Analogue Input (@x), measure 2.5KHz at second analogue input (@y),  simultaneously. 
 

Can I just use On NI USB 6363 BNC connectivity DAQ for this test requirement?  If not, can you suggest any other solution?

 

Best Regards,

 

Marty

0 Kudos
Message 1 of 2
(2,847 Views)

Marty,

 

From the specifications it appears that the 6363 should be able to do that.  I suggest that you generate the signals at a higher level, such as several volts, and then attenuate to the desired levels.  That will give you finer resolution on the generated signals.

 

Lynn