Single High-Impedance voltage source on a DAQ scan (PCI-6355, DAQmx)
I have a DAQ task with 80 channels.  A few of these channels are high-impedance voltage divider sources (~1MOhm or so).  When sampling, I see significant ghosting to the previous channel in the scan.  If I hook up a scope to the test point, I can see a blip and decay that coincides with the overall scan clock.
We have tried to add a 0.1uF cap across the lower leg of the voltage divider:
Source------1M--------<Test Point>--------DAQ
          1M    0.1uF
             GND
(Note:  I cannot change the 1M resistors in this voltage divider)
This helps, but doesn't alleviate the issue.
I have also tried slowing down the rate (just to see what I can do, I still need faster sampling):
  Originally:  1KHz Scan Clock, ~90909 Conversion Clock
  Test:   5Hz Scan Clock, 400Hz Conversion
This also helps, but it's not entirely stable yet.
I am not 100% familiar with the timing; does the MUX switch immediately after the previous measurement?  What is my dwell time for the selected channel?  Can I set that on a per-channel basis, or is it determined solely by the Conversion Clock?
I am considering adding dummy samples on these channels in a group, to increase the dwell.  But if it isn't stable at 2500us of dwell (400Hz conversion clock), then would this help?
(60 Views)
You definitely need to add a unity gain amp or buffer to ensure DAQ is able to sample the signal properly.
Santhosh
Soliton Technologies
(29 Views)
A very large percentage of the time when there is ghosting, you need an op-amp set up in a volage follower configuration before going to your DAQ. This will give the DAQ a low impedance source, allowing the sampling capacitor a place to discharge.