Thermocouple readings not accurate at 0 deg C
We are using an NI-PCI-6032E Multifunction DAQ and an SCC-68 connector block.  We have configured 6 thermocouples on various AI inputs setup in differential mode using the built in CJC.  The CJC is wired to its own AI input per a tech note from NI.   At room temperatures all 6 thermocouples measure within 2 to 3 degrees of each other as well as the CJC. However, when I put each thermocouple in a glass of cold water the temperature reading in the VI is 7 to 9 deg C.  I compare this with a calibrated digital thermometer we have and it is reading -1 to 0.5 deg C.  What gives?
I have been reading a lot of tech notes on the NI website and tried many different scenarios... such as creating a virtual channel using the CJC scale.  Nothing changes.
For testing purposes, I created another VI using a two thermocouples that are setup differently.  One thermocouple is read in using a signal conditioner from Omega Engineering (Super MCJ provides 1mv/deg C) and the other is setup like my 6 thermocouples above (using DAQ Assistan, Type K, built in CJC)  Any insight into this would help.
CJ
(3,350 Views)
Hi CJ,
I would suggest making sure that your input range for the temperature measurements is set to 0-100mV on your PCI-6032E. If the range is on the default -10 to 10 V, you will not get the maximum accuracy. If this is already done, then you may want to calibrate the device within the DAQ Assistant.
Rohan B
(3,312 Views)
Thanks Rohan.  Performing the calibration helped a lot.  I heard that you can send in your PCI-6032E to NI and have them calibrate it. I'm going to inquire about this.
Thanks again BROhan!!
CJ
(3,297 Views)
Glad to hear it, CJ! You can check out Calibration Services for some more calibration information.
Rohan B
(3,279 Views)