10-02-2018 11:48 AM
Hello,
I'm trying to use USB-6002 to measure currents (read voltage over a 0.1 ohm resistor and convert to current). The problem is that input voltage is about 16V (USB-6002's spec is +/- 10V). Would it work if I use differential voltage read (differential voltage across the resistor is small, like 50 mV)?
Or do I need to use voltage divider to lower the voltage to below 10V to make it work even with differential reading.
Thank you for the help.
Solved! Go to Solution.
10-03-2018 10:50 AM - edited 10-03-2018 10:51 AM
Hey,
Even though the difference between the voltages you will be measuring across the resistor is small, the input range of the USB-6002 terminals still needs to be in the range of ±10V.
You will need to lower that voltage with a voltage divider in order to properly use the USB-6002 device.
10-04-2018 06:25 AM - edited 10-04-2018 06:27 AM
Hello,
It's mean you need reading current around 160A, i think you need a current sensor, like:
https://verivolt.com/shop/product/isoblock-i-zf-714?category=13
Regards.
10-04-2018 06:41 AM
I recommend using a Current Sensing Amplifier (such as a TSC103IPT) with your 0.1 Ohm resistor. The amp will convert the differential voltage into a single ended and gain it up to fit within your voltage range. Just apply your scale properly in DAQmx to convert back into Amps.