USB-6009 counter input 5V
Hello everyone,
I am trying to count pulses from an optical sensor.  This sensor outputs 20Vdc for a high value and 0.7Vdc for a low value.  I built a voltage divider that has a total resistance of 12.9Kohms.  As long as I don't connect the output of my voltage divider to the ground and pfi0 (counter) pins on the USB-6009 I get a signal of 4.9Vdc high and 0.175Vdc low which fall nicely into the TTL levels required.  When I make my connections to the USB-6009 my low signal rises to a level of 1.8Vdc and the high level stays the same at 4.9Vdc.  I believe the problem is that when I measure the voltage between the ground pin and the pfi0 (counter) pin of the usb-6009 with no signal wires attached I get a 5Vdc value, and from what I have found this is due to the internal 4.7Kohm pull-up resistor.  My question is, is there anyway I can measure this pulse signal using the USB-6009, I've tried install in 10Kohm pull-down resistor between pfi0 and ground but this doesn't do me any good.  What kind of sensor could handle having a 5Vdc signal present at its output connection?
Labview 8.5
Thank again for the support,
Steven M.
(5,400 Views)
Hi Steven,
To isolate your voltage divider circuit from the internal circuitry of the USB-6009, you can use a voltage follower. This article gives a some additional information on this process with respect to DAQ devices. As for the 5V that you are seeing when there is nothing connected, this is because the pull-up resistor only contributes when the input is floating. When you actually connect a sensor, usually there is a much higher impedance associated with it and the pull-up resistor will have no effect.
Kent
(5,356 Views)
Hello again,
I gave the Op-Amp method a try.  I used a standard LM324 Op-Amp from Radio-Shack.  I have one op-amp on the positive output and one on the negative output.  I have a total impedance of around 17Mohms.  However when i connect the USB-6009 pfi0 input and digital ground my sensor is still getting saturated by the 5volts that is output by the pfi0 line.  I have the op-amps wired in the typical voltage follower method.  My signal line is connected to the positive side of the op-amp and the negative side of the op-amp is connected to the output of the op-amp.
Recap:
My voltage divider works perfect when it is not connected to the USB-6009 Digital Ground & PFI0.
When I connect my divider to the USB-6009 my output levels are saturated twords 5Vdc.
help.
Thank You,
(5,314 Views)
Hi Steve,
You shouldn't need a voltage follower going to ground. I would try the setup again with the voltage follower only going to the digital input.
Kent
(5,292 Views)
Problem Solved Smiley Happy . The voltage follower on just the positive leg worked.  However the the document listed above shows a resitor in series between the op-amp and the digital input.  My first attempt included this resistor (10Kohms) and my behavior was the same as if I did not use the op-amp.  I removed this resistor and connected the output of the op-amp directly to the PFI0 line and TADDA, works perfectly.  Two questions though, the specs of the USB-6009 state that the event counter can take an input from a gated inverter, could I have just connected the output of my voltage divider to the input of a basic IC inverter or maybe a AND gate with one input tied directly to 5Vdc.  Is the event counter just looking for a high impedance source, like that provided by an IC circuit, or is the op-amp required.  And finally when I am at a low state the output of my voltage divider and at PFI0 is 0.7VDC, if I connect a 330ohm resistor in parallel between the PFI0 and GND pins my low voltage goes to 0.3Vdc.  I can reduce this voltage down close to 0Vdc just by reducing the resistor value, however once I get to 10ohms then my HIGH value is effected, up till that point I get a perfect 5.1Vdc HIGH value.  The field wiring guide says I need to use bias resistors when measureing a floating or N.R signal, but they are talking about AI and I am doing a DI.  Is this parrallel resistor acting as my bias resistor and stopping current flow when I am at a LOW state.
I will post a diagram of my working circuit, and of the circuit that did not work later tonight.
Kent, thanks for the directions.
Steve.
(5,287 Views)
Hi Steve,
Glad things are working now. The digital input really only needed some isolation between the external voltage divider and the internal pull-up resistor. Anything which could provide some impedance should suffice. In addition, the resistor that you placed between the input and ground is acting as a pull-down resistor. Generally bias resistors are only used to remove excess charge that builds up on the input.
Kent
(5,266 Views)