Another good troubleshooting step would be to use the test panels in Measurement & Automation Explorer (MAX). Open a test panel for your USB 6259 and use the Analog Output tab to generate a DC voltage.
02-19-2008 02:40 AM
02-19-2008 02:44 AM
02-19-2008 02:55 AM
02-19-2008 11:03 AM
Hi TrH_Linh,
It is beginning to sound like your board may need to be sent in for repair. Before going through that process, let’s try doing a loopback test. The loopback test makes use of your board’s internal channels and so no external wiring will be necessary. Go ahead and disconnect any wires or other connections you have made prior to starting the test, and also close any NI software programs that you may have open.
Directions for Loopback Test:
Purpose: Read an analog output signal on an analog input channel using internal connections.
If you cannot get a reading of 5 volts, then most likely your board needs to be sent in for repair. Please go to http://www.ni.com/support/ and click “Ask an Engineer” to generate a service request and get the repair process started.