USB-6289: +5V pinout doesn't give a signal
My USB-6289 has developed a sudden problem with its +5V DC out. I was checking the signal quality of the +5V signal on my oscilloscope, disconnected, connected a counter (PFI12, ctr0) to the scope, ran a few pulse trains (I am using Igor Pro to control the DAQ), stopped the counter, and reconnected the +5V signal to find that is dead.
I am not sure if I have inadverdantly overloaded the output somehow. The USB-6289 does self-test OK, and the digital and analog output still work fine. It is just the +5V that is not outputting -- is there a software option that may have accidentaly turned it off?
Arun
(3,702 Views)
Here is the DAQ Diagnostic Test 2.0 result. It shows that one of the DAC's (A0) is malfunctioning, but says that the digital i/o is ok -- what does this mean? Can there still be hardware damage?
Selected Device: Dev2
Device Type: USB-6289
Serial Number:
Device Support: UNKNOWN.  The selected device may be newer than your version of the Diagnostic Utility.  Although the tests may appear to run properly, the results could be inaccurate.
NI-DAQmx Version: 9.0 (PASS)
Device Reset: (PASS)
The device was last self-calibrated on: 5-28-2009 3:05:04 AM
Running Calibration ...   (PASS)
-------------------- Testing Counters... --------------------
Testing CTR Buffered Pulse Generation...   (PASS)
Testing CTR Single Pulse Generation...   (PASS)
-------------------- Testing Analog Input... --------------------
Testing AI Ground Reference...   (PASS)
Testing AI Voltage Reference...   (PASS)
Testing AI Finite Sample Clock...   (PASS)
Testing AI Finite Sample Clock Frequency...   (PASS)
-------------------- Testing Analog Output... --------------------
Testing AO Finite Sample Clock...   (PASS)
Testing AO Finite Sample Clock Frequency...   (PASS)
Testing AO Single Point...   (FAIL)
************ Test FAILED for the following reason ************
This test failed to output 9.000000 Volts on Dev2/ao0.  The actual value written was 0.083467.
Testing AO Accuracy...   (FAIL)
This test failed to generate a voltage waveform with an error of less than 0.007621 Volts on Dev2/ao0.  The actual average error was 0.557294 Volts.
-------------------- Testing Digital IO... --------------------
Testing DIO Lines...   (PASS)
Testing DIO Ports...   (PASS)
 ----- DIAGNOSTIC TEST COMPLETE -----
(3,701 Views)
The +5V pin should always be high. Are you sure it's not shorted to GND (or some other low signal), maybe in your DUT board?
If not, I fear the output stage is damaged...
(3,686 Views)
Hey it seems like you card could be broken.
(3,659 Views)