Antonin,

 

I only use SignalExpress.

If you want, i can send SignalExpress project by mail?

 

Regards,

Laurent

0 Kudos
Message 11 of 21
(1,656 Views)

Hi Laurent, 

 

Could you send me a schema of your system : current and resitor wiring.

Indeed your resistor has to be between AI- and the - of your voltage source.

 

Here are two interesting links you could see as well to ensure your wiring :

http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70

http://www.ni.com/white-paper/7114/en

 

Regards,

0 Kudos
Message 12 of 21
(1,653 Views)

Antonin,

 

One setup example (Agilent multimeter is replaced by DAQ USB-6363) :

 

USB_Adaptator_V2.jpg

 

 

Laurent

0 Kudos
Message 13 of 21
(1,651 Views)

Laurent,

 

The NI USB DAQ-6363 is ground-referenced (through your PC), that's why you need to make a conditioning where the shunt resistor will be between the " - " of your power supply and your system.

If the shunt is as on the setup example you gave me, there is a common-mode issue on inputs and a drift in the measurement. To act like your example you would need a NI CompactDAQ configuration with embedded conditioning.

 

Regards,

 

 

0 Kudos
Message 14 of 21
(1,648 Views)

Antonin,

 

 

 

I use DAQ in differential mode, so ground reference is not needed to my mind? Is it false?

 

Regards,

Laurent

0 Kudos
Message 15 of 21
(1,645 Views)

Hi Laurent,

 

Even in differential mode the DAQ needs a ground reference.

 

Let me know,

Cheers

0 Kudos
Message 16 of 21
(1,620 Views)

Antonin,

 

Ok, I will speak with internal HW team to expose connection problem for differential measurement

 

You think it can be source of my problem (measuremnt diff following sample rate) ?

 

 

Regards

Laurent

0 Kudos
Message 17 of 21
(1,616 Views)

Antonin,

 

We are doing power optimisation on embedded system.

We measure current on DCDC converter / LDO input & output by adding serial 0.1ohm shunt. (BGA component)

 

So we can't use advice from your link (http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70)

and use common ground mode

 

I look at “NI CompactDAQ configuration with embedded conditioning” but don’t find what is different compare to USB-6363 ?

 

Is there a solution to use our USB-6363 DAQ in proper way?

 

Thanks

Laurent

0 Kudos
Message 18 of 21
(1,605 Views)

Hi Laurent,

 

Di you try with the configuration I gave you ? What are the results and do you still have similar issues ? 

 

Let me know please,

 

Regards

0 Kudos
Message 19 of 21
(1,591 Views)

 

Antonin,

 

Visio-DAQ.jpg

 

I use this HW setting. I think it fully comply with NI requirements

To simplify i measured directly voltage on shunt that is load (1Ohm/1%)

 

Generator output is sinusoidal signal

 

Measurements:

sample to read=200k, Rate=1M, 1 channel

average = 1.01091mV

 

sample to read=200k, Rate=1M, 2 channels

average = 1.24767mV => ~0.23mV more (+25%) !!!

 

sample to read=20k, Rate=100k, 2 channels

average = 1.01031mV => need to lower sample rate to have good results

 

Regards,

Laurent

 

0 Kudos
Message 20 of 21
(1,585 Views)