11-12-2012 10:02 AM
Antonin,
I only use SignalExpress.
If you want, i can send SignalExpress project by mail?
Regards,
Laurent
11-12-2012 10:18 AM
Hi Laurent,
Could you send me a schema of your system : current and resitor wiring.
Indeed your resistor has to be between AI- and the - of your voltage source.
Here are two interesting links you could see as well to ensure your wiring :
http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70
http://www.ni.com/white-paper/7114/en
Regards,
Antonin G. Associated LabVIEW Developper
National Instruments France
11-12-2012 10:33 AM
Antonin,
One setup example (Agilent multimeter is replaced by DAQ USB-6363) :
Laurent
11-12-2012 11:01 AM - edited 11-12-2012 11:01 AM
Laurent,
The NI USB DAQ-6363 is ground-referenced (through your PC), that's why you need to make a conditioning where the shunt resistor will be between the " - " of your power supply and your system.
If the shunt is as on the setup example you gave me, there is a common-mode issue on inputs and a drift in the measurement. To act like your example you would need a NI CompactDAQ configuration with embedded conditioning.
Regards,
Antonin G. Associated LabVIEW Developper
National Instruments France
11-12-2012 11:09 AM
Antonin,
I use DAQ in differential mode, so ground reference is not needed to my mind? Is it false?
Regards,
Laurent
11-14-2012 03:11 AM
Hi Laurent,
Even in differential mode the DAQ needs a ground reference.
Let me know,
Cheers
Antonin G. Associated LabVIEW Developper
National Instruments France
11-14-2012 08:28 AM
Antonin,
Ok, I will speak with internal HW team to expose connection problem for differential measurement
You think it can be source of my problem (measuremnt diff following sample rate) ?
Regards
Laurent
11-14-2012 11:16 AM
Antonin,
We are doing power optimisation on embedded system.
We measure current on DCDC converter / LDO input & output by adding serial 0.1ohm shunt. (BGA component)
So we can't use advice from your link (http://digital.ni.com/public.nsf/allkb/82508CD693197EA68625629700677B70)
and use common ground mode
I look at “NI CompactDAQ configuration with embedded conditioning” but don’t find what is different compare to USB-6363 ?
Is there a solution to use our USB-6363 DAQ in proper way?
Thanks
Laurent
11-15-2012 07:13 AM
Hi Laurent,
Di you try with the configuration I gave you ? What are the results and do you still have similar issues ?
Let me know please,
Regards
Antonin G. Associated LabVIEW Developper
National Instruments France
11-15-2012 09:33 AM
Antonin,
I use this HW setting. I think it fully comply with NI requirements
To simplify i measured directly voltage on shunt that is load (1Ohm/1%)
Generator output is sinusoidal signal
Measurements:
sample to read=200k, Rate=1M, 1 channel
average = 1.01091mV
sample to read=200k, Rate=1M, 2 channels
average = 1.24767mV => ~0.23mV more (+25%) !!!
sample to read=20k, Rate=100k, 2 channels
average = 1.01031mV => need to lower sample rate to have good results
Regards,
Laurent