03-17-2008 03:05 PM
03-18-2008 06:01 PM
Joe Friedchicken
NI Configuration Based Software Get with your fellow OS users
[ Linux ] [ macOS ]Principal Software Engineer :: Configuration Based Software
Senior Software Engineer :: Multifunction Instruments Applications Group (until May 2018)
Software Engineer :: Measurements RLP Group (until Mar 2014)
Applications Engineer :: High Speed Product Group (until Sep 2008)
03-19-2008 01:17 PM
Hi Joe,
Thanks for getting back to me.
I attach a plot that includes the two waveforms as you asked: the original tissue
response and the input to ground response.
thanks
Keith
03-20-2008 06:23 PM
Joe Friedchicken
03-21-2008 12:58 PM
Hi Joe,
In my application, there is no sensor; I am using acupuncture needles as electrodes
in the tissue. In thitissue measurement above I am just doing acquisition to see if there is any measurable
bioelectric signals. By comparing my tissue response to the open response, I'd wonder
if I'm measuring some response of the op amp to the air or dielectric impendance causing
internal current changes that give the low freq char. that we are seeing? not only noise response?
It's longwinded but in your testing what happens if you ground the -i/p terminal
and leave the + open?
I attach that just test, with wires and with no wires.
Keith
03-24-2008 05:54 PM
Joe Friedchicken
03-26-2008 04:05 PM
Hi Joe,
Thanks for hanging in there with this question.
I attach a diagram and description that should give you a better
picture of the test configuration and what I am trying to test.
I also recreate the original plot of data, and reword my question.
The 30 Hz peaks are curious (in other people there are 80 Hz
and 90 Hz); but not 60Hz. I don't know what this is from. These peaks
have occurred with a desktop and a laptop in 4 different areas of New
England. It is definitely not a local thing.
03-27-2008
09:13 AM
- last edited on
07-22-2025
03:26 PM
by
Content Cleaner
Hi Keith,
Thanks for outlining your wiring. From what I see, I think you could improve your measurements either by including a resistor connection between each AI+ and ground and each AI- and ground, or by not connecting the AI- lines to AIgnd. Your test subjects aren't connected to a building ground, so you're making a connection to a floating source, which means you can use any of the connection configurations in left column of Table 1 of the Field Wiring Guide.
Field Wiring and Noise Considerations for Analog Signals
https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...
Once these changes have been made, I would make another comparison with shorted needles and a full tissue test. From what I know about human tissue, it's highly resistive and slightly capacitive. This means that the tissue looks like an open circuit at DC and low frequencies which is analogous to leaving the inputs open. This also would explain why you see a lot of low-frequency energy in both an open-circuit test and a tissue test.
Joe Friedchicken