05-07-2020 12:52 PM
I'm wanting to use NI9205, specifically the differential channels. I know what channels are tied together for differential purposes (AI0 - AI8, AI1 - AI9, etc.), but my question is how to set this up in FlexLogger. I have the option to set channel AI0 as a differential terminal configuration, but not for AI8. Do I even need to setup Channel AI8 if AI0 is configured as differential?
Thank you.
Solved! Go to Solution.
05-07-2020 06:12 PM
Hey brethigh,
If you configure a channel on 9205 as differential, you do not need to do any additional channel configuration on the corresponding differentially-paired channel.
For example, configuring AI0 on the 9205 with a differential terminal configuration does not require any channel configuration on channel AI8. We will not block you from configuring AI8 as a single-ended measurement if useful to you, but you will not be able to configure a new channel on AI8 as a differential measurement.
For those who see this and may not be familiar with the 9205's differential pairings: http://www.ni.com/pdf/manuals/374188f.pdf#page=15
05-11-2020 06:21 AM
Sounds great, thank you!