10-07-2008 04:26 PM
10-08-2008 03:50 AM
Hi,
Your application sounds very interesting, can I ask what frequency range your signal is coming in at?
For DC voltage application and low frequencies, Pseudo Differential works perfect. For higher frequencies, you should take into consideration that this capacitor is going to look like a short circuit. This may be the reason you are having issues with your PCI-6111.
Best Regards
10-09-2008 08:28 AM
10-09-2008 11:28 AM
Hello again,
Unfortunately the PCI-6111 is limited to Pseudodifferential mode and only has a 12-bit resolution, the USB-6211 is not fast enough to acquire data in the MHz domain. You should be fine using the USB-6211 for initial measurements but when you want to measure higher frequency components the DAQ device will have trouble keeping up with the input and you will loose information.
In order to satisfy the requirements of your application you should look into getting some alternative hardware, look through the products at http://www.ni.com/dataacquisition/.
Contact your local National Instruments branch to your discus options with an Inside Sales Engineer.
Best Regards