Hi,

hope i've posted in the right place. I'm just about ready to purchase a cDAQ-9172 chassis and some 9211 temperature modules or 9219 universal modules. I haven't quite made my mind up yet. I then thought would either of them do the required temperature range that I require. I need to go from -40°C to +200°C and I will be using a K type thermocouple. Does anyone know what the upper and lower limits are with this configuration?

Thanks

Mike

0 Kudos
Message 1 of 3
(3,465 Views)
Solution
Accepted by topic author Mikea01

9211 input specs indicate   Voltage measurement range ..............±80 mV

 

Quick check on Omega web page http://www.omega.com/temperature/z/pdf/z204-206.pdf indicates this combination will easilly cover your range of operation.

 

Message 2 of 3
(3,464 Views)

Thanks for that. I just needed to be sure as its a lot to spend especially if I got it wrong.