Why doesn't the USB-6009 Analog input circuit show the other resistor that appears to go to ground at the analog input on Figure 7 (page 11) of the manual?
Figure 7 (page 11) of the USB-6009 manual shows a 127K resistor feeding a 39K voltage centering bias network prior to the Differential PGA.  This significantly reduces the input signal voltage (good for 10V into a lower voltage ADC, but poor for low level inputs such as Thermocouple).  While planning to reduce this input impedance by a factor of ~10 by paralleling a 14.0K resistor on the 127K (thus upping my input signal significantly), but leaving the 39K, I found another undocumented resistor at the analog input that appears to go to ground.  What is the value and purpose of this other resistor?
(3,095 Views)
Hello EENerd,
This looks to be the 4.7K resister that is going to ground. In order to get specifics of the functionality I urge you to create a service request.
-Travis E
Product Marketer
(3,079 Views)
