Why doesn't the USB-6009 Analog input circuit show the other resistor that appears to go to ground at the analog input on Figure 7 (page 11) of the manual?

Figure 7 (page 11) of the USB-6009 manual shows a 127K resistor feeding a 39K voltage centering bias network prior to the Differential PGA.  This significantly reduces the input signal voltage (good for 10V into a lower voltage ADC, but poor for low level inputs such as Thermocouple).  While planning to reduce this input impedance by a factor of ~10 by paralleling a 14.0K resistor on the 127K (thus upping my input signal significantly), but leaving the 39K, I found another undocumented resistor at the analog input that appears to go to ground.  What is the value and purpose of this other resistor?

0 Kudos
Message 1 of 2
(3,010 Views)

Hello EENerd,

 

This looks to be the 4.7K resister that is going to ground. In order to get specifics of the functionality I urge you to create a service request.

 

Regards,

 

-Travis E

National Instruments
Product Marketer
0 Kudos
Message 2 of 2
(2,994 Views)