Multifunction DAQ

cancel
Showing results for 
Search instead for 
Did you mean: 

Wrong Current with NI9203

Solved!
Go to solution

Hi

I am using the NI 9264 module as a voltage supply and the NI 9203 module to measure the current in my circuit.

In my circuit, I just have a resistor with known resistance. The goal is to calculate the resistance using R=U/I, with U the applied voltage and I the measured current.

 

For a resistor of 10 Ohm, I measure the following currents:

I_1(1V) = 6.6mA, I_2(2V) = 8.6mA, I_3(3V) = 8.6mA, such that I get R_1=151 Ohm, R_2 = 226 Ohm, R_3 = 347 Ohm.

 

For a resistor of 1.5 kOhm, I measure the following currents:

I_4(1V) = 0.6mA, I_5(2V) = 1.2mA, I_6(3V) = 1.8mA, such that I get R_4=1.64 kOhm, R_5 = 1.64 kOhm, R_6 = 1.64 kOhm.

 

I made sure that both modules are connected to the same ground.

 

Does anyone have an idea where this difference (or wrong current) comes from?

Thank you very much!

0 Kudos
Message 1 of 6
(1,354 Views)

NI-9264 is not a power supply.

From the specs:

Current drive for this module is  +-16mA all channels maximum; +-4mA per channel typical

 

0 Kudos
Message 2 of 6
(1,334 Views)

If you're using only one channel, it 9264 can source up to +/-4mA without sag in the voltage. This translates into maximum resistance you can measure up to 2.5k ohm (10V / 4mA).

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 3 of 6
(1,323 Views)

Thank you very much for your answer, I really appreciate that!

Your answer would explain the weird behaviour for the 10 Ohm resistor. However, I still have an offset of 140 Ohm for a 1.5 kOhm resistor.

Do you have an idea where this comes from?

0 Kudos
Message 4 of 6
(1,320 Views)
Solution
Accepted by topic author Yabudabi

NI-9203 has a current shunt internally to measure the current. This value comes to 138 ohms (listed in the datasheet).

 

santo_13_0-1677075234529.png

 

Since 9203 and your resistor are in series, your test voltage is always applied across your resistor + 138 ohm.

 

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 5 of 6
(1,316 Views)

You are right, thanks very much!

0 Kudos
Message 6 of 6
(1,301 Views)