I think this might be a bug, but I would like to know what everyone else thinks. I have multiple analog input cards in my cDAQ and I am reading them at 500Hz. If I add my 9237 card with load cells using the TEDS technology, the samples come into my application at ~1600 samples a second.
My test task below shows continuous samples at 500Hz or 500 S/s, but I get around 1600 S/s. No mater what I change my samle rate to, I still get ~1600 S/s. Do I have something setup wrong, or is this a bug?
Thanks, Doug
Solved! Go to Solution.
09-21-2013 07:29 AM
Here is a snapshot of NI's sample Bridge program showing the timing problem.
09-21-2013 08:39 AM
09-21-2013 09:03 AM
Thanks for the quick response and you are correct that I did not notice the Minimum sampling rate in the specs. The only reason I though it might be a bug was the fact that yesterday, NI tech support could not figure it out and had to do more research. I guess they need to learn the same lesson. In all my years, this is the first card that I used that I ran into the minimum sample rate.