I can not get a reading on an analog input on a PCI-6025E card. My input voltage is on ach0 and I have tied my ground to aignd. What needs to be done to get a analog input reading?
0 Kudos
Message 1 of 4
(3,129 Views)
Bradle95,

The card may very well be in the wrong input mode. The correct mode for the way your signal is wired is Referenced Signal Ended, and the default mode for the card when you install it is Differential.

To change this, go into Measurement and Automation Explorer (MAX), expand the Devices and Interfaces Folder, then expand the [Traditional] NI-DAQ Devices folder. Right-click on PCI-6025E entry and select Properties. Go to the AI tab of the properties window, and on the Mode dropdown menu select Referenced Single Ended.

This will make the board read the signal between the AI channel and AIGND, which is the way you have it wired.

RossC
Message 2 of 4
(3,129 Views)
Great, it works! Thanks for the help.
0 Kudos
Message 3 of 4
(3,129 Views)
Glad I could help out! 😃

RossC
0 Kudos