10-22-2008 01:13 PM
Solved! Go to Solution.
10-23-2008
03:17 AM
- last edited on
07-21-2025
02:53 PM
by
Content Cleaner
Good Morning Exo,
Using a differential connection is definitely a good way to measure an analogue input, as you already understand you halve your channel count when using differential but it will avoid any grounding problems that may occur (eg ground loops).
The following knowledge based article details the different connection types https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider... please refer to "Table 1: analog input connections" for details on how to connect up your circuit. I would recommend using the bias resistors but they are not vital for a differential connection.
Please don't hesitate to contact me if you require any more assistance
Best Regards
10-23-2008 10:53 AM
What is the expected voltage range (with respect to common aka ground) of each of the 'hot' wires?
You may also want to check the specs of your DAQ device for the maximum allowable voltage on the inputs with respect to ground.
-AK2DM
10-23-2008 11:23 AM
