differential voltage levels
Hi!
I have a NI-USB 6212, The analog inputs are speced to -10 to 10 volts. Is it possible to measure a signal which varies between 27 and 27,5 Volts directly. By connecting
27V to CH-
27,5 V to CH+
thus, have a span of 0,5 Volts?
Or do I have do use a voltage divider to get the signal within the spec region and an instrument amplicfication curcuit for better accurcy?
Can the first soulution damage the device?
regards
(3,295 Views)
27V to CH-
27,5 V to CH+
No, you must use a voltage divider. This voltage can't be measured and the card may be damaged. Extracts from page 1 of the specifications :
  • Maximum working voltage for analog inputs
(signal + common mode) ................ ±10.4 V of AI GND
  • Overvoltage protection (AI <0..31>, AI SENSE)
Device on ....................................±30 V for up to two AI pins
Device off ...................................±20 V for up to two AI pins
(3,285 Views)
Thanks for your answer. I guessed that was the case.
Somehow I just figured since the the ch- was 27V and the ch+ was 27,5 the actual potential was 0,5V, well within the spec, which was going to safe some curcuits. Oh well I have to solve it the, well not so, hard way.