measuring output state on USB-6525
Hi,  is it possible to verify the state of the SSRs in a USB-6525 using a multimeter.  I connected a meter to measure the resistance across the output connections  P0.0A and P0.0B.  I wanted to use the self-test routines to verify that the SSRs were closing.  As expected there was an open circuit when the TTL was 0,  but the circuit did not appear to close when I changed the TTL to 1.  Do I need to connect a power source and load to be able to see the SSRs closing?  Isn't that what a multimeter is?
Thanks,  Nick
(3,331 Views)
Accepted by NJGW
Nick,
Multimeter will not provide the voltage source and load you need.  See this link.  http://digital.ni.com/public.nsf/allkb/B0FF01B2AFFA52FC862573A2005A9570
(3,322 Views)
Hey Nick,
As the KB Centerbolt posted explains, we need an external power source to provide power to the device under test.  The relay module does not provide any power; it simply switches on or off.  The DMM you're using provides a small bias voltage in order to measure the response current, which is then divided out (IR=V) to find the resistance.  The amount of power sourced from the DMM is very very small and is not enough to overcome the internal effects of the semiconductor circuitry.
-John Sullivan
Problem Solver
(3,312 Views)
