Hi, is it possible to verify the state of the SSRs in a USB-6525 using a multimeter. I connected a meter to measure the resistance across the output connections P0.0A and P0.0B. I wanted to use the self-test routines to verify that the SSRs were closing. As expected there was an open circuit when the TTL was 0, but the circuit did not appear to close when I changed the TTL to 1. Do I need to connect a power source and load to be able to see the SSRs closing? Isn't that what a multimeter is?
Thanks, Nick
Solved! Go to Solution.
01-08-2009 07:51 AM
Nick,
Multimeter will not provide the voltage source and load you need. See this link. http://digital.ni.com/public.nsf/allkb/B0FF01B2AFFA52FC862573A2005A9570
01-08-2009 01:01 PM
Hey Nick,
As the KB Centerbolt posted explains, we need an external power source to provide power to the device under test. The relay module does not provide any power; it simply switches on or off. The DMM you're using provides a small bias voltage in order to measure the response current, which is then divided out (IR=V) to find the resistance. The amount of power sourced from the DMM is very very small and is not enough to overcome the internal effects of the semiconductor circuitry.