Multifunction DAQ
pcie 6363 - voltage drift in SE recording
I am recording the output of a function generator (~15mV sinusoids at ~3000 Hz). I am using a BNC to direct the signal out of the function generator and into the BNC2090A and recording the analog signal in SE mode.  While the data starts out being close to 0 in diff mode (fig 1), it drifts towards -10V as soon as i shift the recording to SE mode (fig2 ). Finally after ~ 5 minutes, the voltage eraches its minimal value and levels off (fig 3)
Whats happening?
Board: PCIe 6363
PC: Dell optiplex running Windows 7
DAQ Interface: BNC 2090A
Only AI0 is connected. Every other BNC slot is empty
I checked the output of the function generator using an oscilloscope - it is pretty stable over many hours.
Balaji
(4,627 Views)
When you switch to single ended mode, how are you connecting the gnd? Out sounds like you have a floating signal.
(4,610 Views)
Thanks for the reply.
I am recording in RSE mode(if that is what you are asking)
(4,603 Views)
You already said that. I'm asking about the physical connection changes that you changed.
https://www.ni.com/en/shop/data-acquisition/measurement-fundamentals/field-wiring-and-noise-consider...
(4,596 Views)
What Dennis is trying to get at is that if you are using RSE configuration, you need the ground of your signal to be connected to AI GND.
GCentral
(4,588 Views)