 SBD_Testing
			SBD_Testing
			01-08-2014 03:12 PM
I am trying to setup a USB-6210 to read pressure transducers (specifically: http://www.omega.com/pptst/PX309-100mv.html), with labview2013.
Its a fairly simple task, and it should not be this hard...
I have used a similar system to take similar measurements in the past, but its been a while.
I'm not sure if its just a wiring issue or incorrect labview setup (program), but I can't seem to measure any voltages that are remotely close to the transducer's calibration sheet.
Thoughts?
Solved! Go to Solution.
 Knoxberry
			Knoxberry
			01-09-2014 04:32 PM
Do you have any way to verify that the voltages you are reading from the pressure transducer are correct? You could use a DMM if you have one around. How do you have your system wired? What voltage are you expecting and what voltage are you measuring for a particular pressure? Is the voltage off by a constant amount for different pressures? I'd need more information to help you troubleshoot.
01-09-2014 05:24 PM
Turns out my power supply wasnt outputting properly. Switched to a different power supply and everything works as it should.