Hi Zhe,
Before taking a look at the frequency reading on the FFT, I will like to know in the time domain: what is the average reading when a force is applied? What happens if nothing is connected to the 9237 do you still read 100, 200… on the FFT? When you do the FFT what is the value (dB or Volts) of those frequencies? How fast are you expecting your signal to change? Can you explain a little bit more why are you taking the FFT of your signal I would like to know the information are you trying to get from it?
Since the sensor is placed in a vibration isolation table; like you said, those frequencies are more likely to be noise either in the cable/connections or in your Lab. What we need to do is first identify if it is noise and the source of this noise. Some suggestion I have for you are changing the slot in the cDAQ-9172, if it’s possible move the cDAQ to a different place on your lab or a different room, change the sampling rate to make sure we are not oversampling, disconnect any other module from the 9172 when doing this troubleshooting. While I’m waiting for you to answer those questions I will suggest taking a look at: Measuring Strain with Strain Gauges.
I hope this helps
10-25-2007 03:37 AM
10-26-2007 12:54 PM
10-31-2007 04:35 AM
11-01-2007 10:32 AM