sample rate / resolution relationship for USB-6351
Do the AI resolution or AI accuracy on the USB-6351 vary at all with the sample rate?  For example, the datasheet for the 4070 DMM contains this chart of the sample rate/resolution relationship:
However, I don't see anything like this in the datasheet for the USB-6351.  I see where the absolute accuracy depends on the voltage range, but nothing tied to sample rate.  Does that mean I will get the same resolution (16-bit) and accuracy regardless of whether I am measuring a single channel at 10kS/s or 2 channels at 500kS/s each (1MS/s aggregate)?  Thanks.
(3,457 Views)
Hi Reedayoungblood,
From the datasheet, the USB-6351 can sample one channel at a max rate of 1.25MS/s with 16 bit resolution. So as long as your aggregate does not exceed 1.25MS/s you will not get losses in accuracy or resolution. I hope that helps.
(3,434 Views)
The FlexDMM has a sort of on-board averaging, so sampling for a longer duration will give a higher resolution.
The 6351 on the other hand uses a SAR ADC which is fixed at 16 bits of resolution.  However... your question about accuracy isn't as straightforward.  Changing the sample rate if you are acquiring from a single channel shouldn't affect accuracy.  When acquiring on multiple channels however, settling error can become a factor (there is a chart on page 2 of the specifications).  You need to allow enough time for the voltage on the ADC to settle to within the desired accuracy if you are acquiring on multiple channels.
Best Regards,
John Passiak