scc-sg24 and gain problem

Hi to everyone,
I'm using a SCC-68 connector with a scc-sg24 module to acquire a load cell with this features:
-model FN3060
-Supply Voltage 10 Vdc
-F.S.Output 2mV/V

The problem is that the voltage value acquired in MAX is the same that I measure with multimeter as input of scc-sg24 module.
In the attached file you can see an example: the amplitude is about 1mV, but that I expect is 100mV if the gain of the module is 100.
Any suggestion about this problem ?
Thanks.
Bye


Message Edited by Garraty on 02-04-2008 12:21 PM

Message Edited by Garraty on 02-04-2008 12:21 PM
0 Kudos
Message 1 of 2
(2,887 Views)

Here is a blurb from the manual.

               If you have not configured the SCC-SG0

X or SCC-SG24 as an accessory

               for the E Series DAQ device as described in the

Calibration section, or if

               you are using the Macintosh operating system, you must divide voltage

               measurements by 100 before converting them to units of strain. This

               calculation accounts for the gain of 100 applied by each SCC-SG0

X or

               SCC-SG24 channel.

This implies that under certain conditions LabView will automatically compensate for the gain of the SG module.