Hello wanfaz,
The only limitations that Signal Express Limited should impose on you is that only a single log file can be recorded and loaded at once and some of the additional signal processing and analysis steps are not available. There should be no limitation on the length of time you are able to acquire for. That being said, can you be more specific about what is going wrong?
What exactly is occurring?
You asked about the buffer size. Are you receiving an error?
Are you acquiring analog or digital data?
Is your acquisition continuous or finite?
With more information we can try to get you back on your feet!
Regards,
01-29-2008 11:02 PM
01-30-2008 05:55 PM
01-30-2008 11:02 PM
01-31-2008 10:21 AM
Hi wanfaz, Ryan,
You need to keep in mind the differences between the N Samples and Continuous sampling methods. When you use N Samples the Samples to Read entry corresponds to how man samples you want to read. When using Continuous sample however this value controls the size of your buffer. The Data Discontinuity Error usually occurs when you overwrite the buffer before you read the data. Can you give me some more detail on what is going on?
Specifically:
What is the actual DAQmx error that is thrown? The Data Discontinuity error is thrown by Signal Express, however there is usually a DAQmx error that leads to this. You can find this error by looking at the DAQmx step and seeing if there is an error present.
You mentioned in your first post that you were using recommended values for Samples to Read and Sample Rate. This is usually a 1:10 ratio respectively. In your latest post you say you are reading at 10 Hz and 100 samples. Which one are you actually using?
I am running a simple test as we speak with your settings and I will see if I can recreate the error.