temperature output from RTD is not correct
Hi..everybody.
I am new for Labview, I need to get the temp-value from RTD(PT100) which connect with SCXI-1328 and SCXI1121 (amplifier).
I try to use traditional DAQ --> choose RTD (with temp measurement). and the result is far away from actual tempurature.
I try agian by setting DAQ --> voltage and then use RTD converting (I devide the voltage input with 100 before go to RTD converting because I set gain=100 at SCXI 1121).
Both of them gave me the wrong result. I don't know that which step is wrong.
Could anybody help me please.
Thank you very much in advance
(3,058 Views)
Hello Nok,
Let me explain first how your RTD measures temperature and then I will talk about how the software displays the temperature.
An RTD is a Resistive-Temperature-Detector. It looks like you have platinum RTD which has a nominal resistance of 100Ohms at 0 degrees Celsius. As the temperature of the RTD changes, so does its resistance. The SCXI-1121 has excitation that is used to pass a constant current through your RTD. You can choose to excite your RTD with 0.15mA of current. This 0.15mA current will pass through the RTD. As the resistance of the RTD changes due to a change in temperature, the voltage across the RTD will change, but the current passing through the RTD will remain constant at 0.15mA. Therefore, if you were to measure the voltage ac
ross the RTD at 0 degrees Celsius, your RTD would have a resistance of 100Ohms which yields: 0.15mA*100Ohms = 15mV
You do not have to do any dividing in software, because the hardware knows to divide by the value of the gain you have applied.
In order to get this voltage to be accurately converted to a temperature, I recommend that you use a Virtual Channel. This will allow you to configure the excitation and parameters for your RTD and will also return the measurement in degrees rather than volts. To create a virtual channel, open Measurement and Automation and then right-click on the Data Neighborhood folder and select Create New...
Here is a link to a tutorial with more information and links: http://zone.ni.com/devzone/conceptd.nsf/webmain/D9BAA2AA872A6B5F86256D730079E3E6?opendocument
Bill B.
(3,058 Views)
Thanks for ur help .. Bill B.
One question that I still have some doult is about the gain that the hardware can detect itself. U mean DAQ card can know that which value of gain that is set on SCXI 1121 ?
Because of my program have to connect to many sensors so I prefer to use only get voltage from SCXI-1121 and then voltage go throught different calculation depend on user. (because I can use common acquirsition).
In this case I have to pay attention on Gain of SCXI ?
Thank again for your great help.
Nopporn