06-16-2008 08:16 AM
Can you rephrase or expand? I tried reviewing the thread and I'm not sure I know what measurements need to be kept separate. I thought the main problem was to to synchronize tasks, 1 of which is analog input and the other of which is counter-based encoder position measurement.
-Kevin P.
06-16-2008 09:51 AM
06-16-2008 02:12 PM
Not sure you're going to be able to get there from here. You can only have 1 hardware-timed AI task running at a time on an M-series board -- and using the change detection event counts as hardware-timed. The only way to have 2 independent and simultaneous AI tasks is by using an on-demand software-timed scheme.
Also, you can similarly only have 1 digital input change-detection task at a time.
Best potential workaround I can think of:
A. Create 1 DI task that uses change detection. Wire both channels of both encoders to digital inputs on port 0 and be sensitive to both rising and falling edges.
B. Create 1 AI task that uses the change detect event as the sample clock. Include both AI channels of interest in the 1 common task.
C. Create 2 counter position measurement tasks for the 2 encoders that use the change detect event as the sample clock.
D. Sort out the extra data with some software post-processing. Each sample taken will have been ultimately caused by either encoder 1 or encoder 2. By tracking the changes in encoder value or DI bit states, you can determine which encoder caused each sample. Then you can pay attention to the appropriate AI channel at that sample #.
E. Simple code will get you about 95% there. Little gotchas that remain include: handling rare cases where both encoders changed simultaneously, and handling the very first sample where you can't easily determine which bit changed. Also, be sure to start the AI and CTR position tasks before starting the DI task. This way, all the tasks depending on change detect events will be armed and waiting before any changes are detected, thus ensuring that their samples will be in sync.
-Kevin P.
06-17-2008 03:18 AM
Thanks Kevin. My backup plan is to trigger both the analog and the counter tasks on the edges of each encoder (which is connected through an LS interface circuit).
This will give me hardware timing, but I will have the problem that the tasks are out of sync when they start as each startTask() call takes some time. However this can *almost* be solved by starting the CI task before the AI task, then checking how many samples were read on the analog and counter task the very first time (I have a while loop where both buffers are read). I then remove any extra counter samples as the CI task began before the AI task, and those samples are from before the AI task started.
There will still be a small sync error though, because the two sequentital reads will be separated a bit in time. However as the frequency is moderate (5 kHz), only one or two encoder ticks passes between the reads() and that is an acceptable error for this application. Thanks for all help gain!
/ Roland