NI LabVIEW,CVI,数据采集等产品讨论区
取消
显示结果
搜索替代
您的意思是:
PCI6602的counter每1ms采一个数要设置什么参数(测速)
Measurement Method:2 Counters(High Frequency)
帖子被ThomasYoung在 01-31-2010 07:25 AM
时编辑过了
Thomas | Xi'an China
LabVIEW 7.1 | LabVIEW 2018
DAQmx | CompactRIO | myRIO
微信公众号: 钟博士LabVIEW工作室
1 条消息(共 9 条)
5,585 次查看
这个问题我也想知道...
怎样设置参数使PCI6602的采样时间为1ms?
或者怎样使用硬件定时为1ms?
0 项奖励
2 条消息(共 9 条)
5,556 次查看
You can configure your task as image below
it will buffer your result every 0.001 second. Check it in the example finder:
hardware input and output->DAQmx->counter measurement->Digital frequence->Meas Dig Frequency-Buffered-Cont-High Freq 2 Ctr.vi
And for ok-forward:
You can check in your example finder, under DAQmx->counter, every example that has buffered in its name is hardware timed counter measurement.
You can refer to the examples.
Best Regards
Hu Yu
3 条消息(共 9 条)
5,546 次查看
High Freq 2 Ctr
转速是250r/min,≈2083Hz
0.0001s时,有error-200302,Measurements:Signal being measured is slower than the specified measurement time.
0.001s时,测出的数据错误,不是2000Hz就是3000Hz
Low Freq 1 Ctr
转速是250r/min,≈2083Hz
error-200474,the specified timeout expired.
0 项奖励
4 条消息(共 9 条)
5,541 次查看
Well, if you take a look at the frequence measurement basic, you should know why these error would happen.
SO:
1. You can never use a 10K read frequence to read a 2K signal, there is no result for you to read
2. If you want your result to be correct, use a measure period longer enough than your signal period, at least 5 times longer. You can image to use a 1K sample rate acquire 2K square wave, any disturbance will cause signal loose, not to speak of the rotation rate of a motor, its quite unstable.
3. This is because, I think, you are using a 20M timebase, you can caculate how many counts your counter should have. I suggest you input the frequence range, or use a 100K timebase
Message Edited by NIhuyu on 02-01-2010 06:51 PM
Best Regards
Hu Yu
5 条消息(共 9 条)
5,538 次查看
我需要测量的频率范围是0~3000r/min,应该使用哪种方法?
多大频率算是高频?
0 项奖励
6 条消息(共 9 条)
5,532 次查看
Just as I said before, use a 1 counter frequency should work, but I think you should change your timebase to 100KHz, by property node or by specifiying the frequency range, others may not work properly,
FOR the high or low frequency
There isn't a specific bandwidth that is defined for use with either the one-counter or two-counter method. However, located in the NI-DAQmx help file there is a section detailing the quantization errors that arise when using either method with a variety of input frequencies. By referring to this information you should be able to determine which method will produce the smallest error in your application and hence which method will be preferable. This help file also explains the errors associated with the large-range two-counter method.
To locate the help file, you should select Start»All Programs»National Instruments»NI-DAQ»NI-DAQmx Help, and search for 'quantization error'.
I hope this helps. And sometimes I think you can move on to try some more settings, because it will cost you nothing, nothing will damage.
Best Regards
Hu Yu
0 项奖励
7 条消息(共 9 条)
5,516 次查看
轴是转的,用High Freq 2 Ctr可以测量,但用Low Freq 1 Ctr测不出数据。频率范围正确
because the specified timeout expired
帖子被ok~forward在 02-02-2010 02:01 AM
时编辑过了
0 项奖励
8 条消息(共 9 条)
5,492 次查看
What's your Low Freq 1 Ctr timebase? You checked it with the DAQmx property node?
According to the image below, the 100KHz timebase can measure from 10 to 10KHz, there's no reason you can not test the FREQ
Best Regards
Hu Yu
0 项奖励
9 条消息(共 9 条)
5,478 次查看