NI TestStand
Continue sequence even if one instrument is missing?
Is there a way to take care of or avoid the following error message ””VI_ERROR_RSRC_NFOUND   BFFF0011h   Insufficient location information or the device or resource is not present in the system.”?
I have two deployment system, lets call them A and B. In system A there is two R&S SFUs and in system B just one R&S SFU.  I want to use the same sequence file on the both system. This is due to easier maintainence and flexibility.
The problem occurs in system B when the sequence file tries to open the second SFU, that is missing.  I expect the error message but how can i take care of it, or prevent it,  in order to continue the sequence?
Test Stand reports the following error "-17502; System Level Exception."
The MAX output is:
> 973.  viOpen (0x0D39F1B0, "TCPIP::192.168.1.25::INST0:...", 0, 0, 0x00000000)
> Process ID: 0x00000DAC         Thread ID: 0x00000D0C
> Start Time: 14:47:27.541       Call Duration 00:00:02.000
> Status: 0xBFFF0011 (VI_ERROR_RSRC_NFOUND)
Attached is a very short sequence file showing the problem .
The dll can be downloaded from:
http://www2.rohde-schwarz.com/en/products/test_and_measurement/product_categories/video_tv/digital_m...
I am very thankfull for any input!
Mange
(3,431 Views)
Accepted by Mange
Try setting the Run option - Ignore Runtime Errors for the Open step this will mean your instrument handle will be zero so you can use this as a precondition on the other steps to skip if its zero.
Regards
Ray
Regards
Ray Farmer
(3,412 Views)
Many thanks Ray ,
You solved my problem! This is an accepted solution.
Magnus "Mange" Lindell