Hi Ana10,
Are you using this digitizer with NI VideoMaster? if not you should probably post this in the Digitizer forums. That said I would suggest using the self calibrate function in MAX for this device rather than the LabVIEW API method just so that you can rule out any errors in correctly configuring the digitizer for self cal in LabVIEW. Also you should ensure that all inputs are disconnected before performing a self cal. If this still results in an error in calibration you could refer to the following document or arrange to return the digitizer to NI for external calibration.
http://www.ni.com/pdf/manuals/370328e.pdf
Hope this helps,
Nick