Hi,
I would like to ask about the setup of your application. Are you trying to run a VI on your Windows PC which acquires data remotely from the PXI-6070E? If so, RDA is the proper method. However, if you run the VI on the PXI-8146RT target instead, you don't need to use RDA, you can simply use DAQmx. The reason I bring this up is because RDA is only supported by Traditional DAQ and not DAQmx.
As I understand it, you can see all your devices under
Remote Systems » OA48F3 » PXI-System » Chassis. From your screenshots it appears that you have Traditional DAQ 7.3.0 installed on the PXI-RT controller (
Remote Systems » OA48F3 » Software). However, what is the version of Traditional DAQ installed in your local PC (check under
My System » Software)?
Both these versions have to be the same. If not, you can right-click on
Remote Systems » OA48F3 » Software and select
Install Software and install the current version onto the controller.
Also, we need to ensure that your PXI-6070E device is working properly. Best way to check this is to take all the other devices out and only plug in the PXI-6070E. Also, try different slots to make sure none of the slots are bad.
Please try these things and post a response. We will do our best to help you out with this.
Have a great day!
Anu Saha
Academic Product Marketing Engineer
National Instruments