Howdy altran,
Have you installed NI-VISA onto your PXI-8196 RT controller? I assume you've installed NI-VISA onto your Windows PC (If not, you can download NI-VISA 4.3 here. ) But in addition to having VISA installed on your Windows PC, you also need to install NI-VISA onto your Real-Time controller. To verify that you have VISA installed onto your RT controller, follow these steps:
1. Open Measurement & Automation Explorer (MAX)
2. Navigate to your 8196 under Remote Systems
3. Expand the Software tree beneath your 8196 system
4. Verify that NI-VISA is listed.
If you don't see NI-VISA listed, then you can install it onto the RT controller by right-clicking on Software and select Add/Remove Software.
Instead of referencing your Sorensen power supply through a TCP/IP VISA Resource in MAX (aka a VISA alias), you might also consider accessing the power supply in your VI directly by using this syntax: TCPIP::xxx.xxx.xxx.xxx::INSTR, where the x's represent the power supply's IP address.