PXI
How do I change the configuration channels for Analog Outputs on a PXI-6704?
I can't seem to figure out how to access the configuration channels for Analog Outputs (Channel 33 and 35) for the PXI-6704.
This piece of test equipment is out of tolerance, and needs to be calibrated.
I've followed the calibration procedure http://www.ni.com/pdf/manuals/374081b.pdf, but am getting hung up on the step 8 where it says to repeat steps 4-7 while changing _cal_ao_current_offset. I can't figure out how to open this channel as it is not listed as a physical channel for step 5.
I'm using Labview 7.1
(3,444 Views)
Accepted by Green_Tim
Hi! What you need to do is add "Internal Channels" to your channel list so you can select the _cal_ao_current_offset or _cal_ao_voltage_offset.  This is done by right clicking the channel drop down list and select "I/O Name Filtering...".  There will be a check box that says "Internal Channels" and will provide you all the available internal channels to input into your Create Channel vi.  I attached some pictures illustrating how to do this. Cheers!
Jordan F
(3,426 Views)
