We have a test system with 7966R cards and are getting occasional error messages, see attached.
Has anybody had similar problems?
Solved! Go to Solution.
01-02-2024 09:41 AM
Is this a new system under development or an existing tester that is starting to cause issues?
IO module may not be securely connected to the 7966 causing intermittent errors.
The two errors are different but related to some kind of timeout. Bitfile load time should not be too long. Need more info on the circumstances causing the error.
01-02-2024 10:34 AM
The test system has been in use for a few months and has now started throwing errors. We suspect a poor connection between the PXIe-7966R and the backplane in the PXI rack. We will try removing and reinserting the card to see if it cures the communication problem.
01-02-2024 10:50 AM
@sandals18 wrote:
The test system has been in use for a few months and has now started throwing errors. We suspect a poor connection between the PXIe-7966R and the backplane in the PXI rack. We will try removing and reinserting the card to see if it cures the communication problem.
I would also check the adapter module, the card connected to the 7966 on the front end that provides I/O.
01-03-2024 04:23 AM
We have just disconnected the adapter module and unplugged and refitted the 7966R. problem solved!
Apparently there was a weak electrical connection somewhere which has been improved by unplugging and replugging.