Is the PXI-67xx analog output reference (AO GND) isolated or tied to chassis/facility ground?

Solved!
Go to solution

I'm trying to simulate pressure transducers with an output of 3mV/V (excitation is +10V, so full range should be 0mV to 30mV).  I'm looking for a high channel-count analog output PXI device (possibly PXI-6723) that does not have the low side of the analog output channels tied to facility ground (differential AO channels).  The plan was to use a voltage divider circuit to scale the +/-10V output of the AO card down to +/-50mV.  I'm using a 16-channel, 12-bit AO card that has all of the AO commons tied to the PXI chassis ground and facility ground.  The DC part of the signal does what I want it to, but there is approx 80-100mVpp of noise riding on it (from the ground connection).  I also tried a PXI-6115 card which we had on-hand, and its AO reference is tied hard to chassis as well --- same result with the ground noise.  Is the PXI-6723 AO GND isolated from chassis/facility ground?  Even if all of the AO GNDs are shared on the board, that solution should be much better than having a noisy ground reference. 

 

Any ideas?  Alternate solutions to simulating a floating, millivolt output device?

 

Thanks!

0 Kudos
Message 1 of 3
(4,011 Views)
Solution
Accepted by tmtowtdi

The AO reference of the 6723 is tied to ground; the product pages the search on channel to channel isolation analog output, only a handful of AO devices have an isolated ground (the trade of is a low channel count).

 

-Ciao

Message 2 of 3
(3,982 Views)

Thanks for your quick response!  I appreciate it

0 Kudos
Message 3 of 3