PXI
Is there a common set of hardware that one uses with Calibration Executive?
Which hardware would be appropriate to use with Calibration Executive to calibrate a system containing solely PXI-4462 Dynamic Signal Acquisition Cards and PXI-6653 Timing and Multichassis Synchronization Modules?
Is Calibration Executive capable of generating the NIST traceable calibration sheets?
Can Calibration Executive provied as received before calibration data as well as post calibration data?
In other systems we use mostly PXI and SCXI cards such as:
PXI-4071 7½-Digit PXI Digital Multimeter
PXI-4110 Triple-Output Programmable DC Power Supply
PXI-5105 60 MS/s, 12-Bit, 8-Channel High-Density Digitizer
PXI-4461,62,72 24-Bit, 102.4 kS/s, 204.8 kS/s Dynamic Signal Acquisition
PXI-6254 16-Bit, 1 MS/s (Multichannel), 1.25 MS/s (1-Channel), 32 Analog Inputs
PXI-6508,11,12,13,14,15,21 Various Digital I/O
PXI-6723,33 High-Speed Analog Output -- 13-Bit, 16-Bit
PXI-6651,52,53,82 Timing and Multichassis Synchronization Module
SCXI-1100 32-Channel, ±10 V Analog Input Module
SCXI-1102 32-Channel Thermocouple Input Module
SCXI-1520 8-Channel Universal Strain Gage Input Module
Kevin.
Subliminal:  I could save the world if I could put a Tab inside a Cluster.
Message Edited by kmcdevitt on 03-30-2009 03:47 PM
(3,191 Views)
Lookup in the Help files here   collect all named References (plus add. cable sets) and wipe out doubles or the lower spec devices 😉
Short guess:
Fluke/Wavetek 5720A,  9500 , HPAG 3458A .....
OR look at your applications, find out what you need (all functions of a card, or just DC Voltage in one range, 20ppm or 1%) and maybe reduce it to a good function generator and a calibrated HP3458 (a real HP masterpiece IMHO)
ALWAYS needed, rarely found in the lists: skilled & trained operators 😉