hi,
while trying to calibrate a PXI-4060 DMM module (procedure version 2.0) under calibration executive 2.0 following error is screened and the calibration fails:
Error executing substep 'Post'. The following SQL command failed: 'SELECT RESISTANCE.ID, RESISTANCE.Resolution, RESISTANCE.Delay, RESISTANCE.AutoZero, RESISTANCE.Range, RESISTANCE.Test_Point, RESISTANCE.Error_24_Hour, RESISTANCE.Offset_24_Hour, RESISTANCE.Error_1_Year, RESISTANCE.Offset_1_Year, RESISTANCE.Error_Temp, RESISTANCE.Offset_Temp, RESISTANCE.TempFactor, RESISTANCE.Lower_Limit_24_Hour, RESISTANCE.Upper_Limit_24_Hour, RESISTANCE.Lower_Limit_1_Year, RESISTANCE.Upper_Limit_1_Year FROM RESISTANCE...'
Native error code -2147217904 0x80040e10
Mi
crosoft OLE DB Provider for ODBC Drivers:
[Microsoft][ODBC Microsoft Access Driver] Too few parameters. Expected 8.
NI-DAQ 6.9, IVI-engine 1.83 and Test-Stand 1.0.3 are installed on the system. Even the calibration with an older version (1.1) of the procedure fails (ividmm.dll error message).
When using calibration executive with E-Series modules the calibration can be performed without any error messages.
How can we fix this problem? We appreciate your help.
Best regards,
Dipl.-Ing. Philip M. Fleischmann - CTO
esz Elektronik-Service GmbH
Max-Planck-Str. 16
D-82223 Eichenau
Phone: +49-8141-88887-39
Fax: +49-8141-88887-77
e-Mail: p.fleischmann@esz-gmbh.de
http://www.calibration-support.com