02-28-2019 01:28 AM
PXI 4130 SMU no aux output during calibration.
03-01-2019 09:20 AM
JVmaster
Can you describe a little better what you are experiencing?
Are you following the Calibration procedure described in this PDF (http://www.ni.com/pdf/manuals/374753e.pdf)?
03-01-2019 06:44 PM
the PXI 4130 is working at the production, when the unit submit for calibration got failed No aux power output during verification..
We have 2 problem on our PXI 4130.
1. failed and accuracy
2. no aux power output.
03-01-2019 06:54 PM
attached is the validation result of units failed in calibration