PXI

cancel
Showing results for 
Search instead for 
Did you mean: 

PXI-6704 voltage output latency

 

I have on old PXI-6704 and I want to generate 1 channel, 1 sample voltage output.

 

Everithng  is working as expected exept for the output latency. I can see all samples are generated (the update rate is 5ms bigger than the max of 1.8ms), but I can messure a delay of aprox. 50ms from the moment of  DAQmx Write and the Voltage generated.

 

Is there any method to configure the card to reduce the FIFO.  I alredy tried a lot of samples from people with similar problems but their solution did not worked on my PXI-6704.

 

Any suggestions will be mutch appreciated,

 

Thanks.

0 Kudos
Message 1 of 3
(4,716 Views)

 

Hello Valic,

 

I see that this instrument is software time, so maybe the OS is affecting your operation as it is non-deterministic.

 

I looked up the supported properties on your PXI 6704 and it seems it supports the "DAQmxSetBufOutputBufSize" and "  DAQmxSetAOUseOnlyOnBrdMem" properties for the output buffer, have you tried setting that to first one to 0 or set to use only on board memory?

 

You can look at the supported properties and descriptions here:

 

http://zone.ni.com/reference/en-XX/help/370471W-01/cdaqmxsupp/pxi-6704/

 

 

Regards,

 

David M.

National Instruments 

 

 

0 Kudos
Message 2 of 3
(4,698 Views)

Thank you David for your response,

 

I tried to set to 0 the Output.BufSize but get on unsupported error 200452.

Also when I set to true AO. UseOnlyOnBrdMem I get error 200077 indicating that I can only set to false this property.

My OS is LabView Real-Time 8.5 and I saw I can update DAQmx to version 9.3 and hope to have better results.

Or maybe I can access the card directly by VISA protocol which is something I have no experience (registers, comamnds, etc for this card).

0 Kudos
Message 3 of 3
(4,677 Views)