Ennio,
The PXI-1000b, PXI-1006, PXI-1010 and PXI-1011 chassis are designed to operate on a bench or in an instrument rack. They can operate at temperatures up to 50ºC, but we recommend to install them in places where there is air flow for the cooling system of the PXI chassis to work well.
Unfortunately I cannot tell you what could go wrong if you operate your PXI chassis under conditions that are not recommended. Any component of the chassis, or even a board the chassis, could get damaged under this conditions.
The following is a paragraph that I extracted from the PXI-1000B manual.
It applies to the chassis mentioned above, but I recommend you to consu
lt the manual for your specific chassis.
"Apertures in the rear and along both sides of the mainframe facilitate power supply and module cooling. Air enters through filters and fan inlets located
in the lower rear of the mainframe and exits through the upper sections on both sides and through the rear.
Place your PXI-1000B on a bench top or in an instrument rack so that the fans (air inlets) and the air outlet apertures along both sides of the mainframe
have adequate ventilation. Keep other equipment a minimum of 3.0 in. (76.2 mm) away from the air inlets and outlets.
Install your mainframe so that you can easily access the rear panel. This simplifies the replacement of the air filters or power supply/fan assembly, if necessary."
Hope this helps,
Claudia Lorente
Applications Engineer
National Instruments