PXI
PXIE-4163 error code 1073364992
HI
I have a problem during measurements of one of the channels one the PXI-4163, when measuring I get the warning 1073364992, and return nan values.
Am using python where I am measuring several channels in a loop, using the measure multiple, but this always fails with this particular channel. More information is that this channels is configured to be a current source with a voltage compliance. The channel has its output on during the measurements.
(1,584 Views)
Could you share the content of the exception?
Santhosh
Soliton Technologies
(1,543 Views)
warning DriverWarning:warning 1073364992 occurred measurement over range
(1,497 Views)
I have finally found the cause of the error, and fixed it,
When it is happening when current_level and current_level_limit are equal,  and the load is able consume just above the current_level, then we get that measurement range error. The solution is always to have the current_level_limit higher than the current_level.  The error does not occur when the current_level is slightly below the current_level_limit.
(1,396 Views)