10-12-2012 08:43 AM
Can anyone tell me if the PXIe 4300 has separate grounds for each channel? I need to connect several signals to it in my application without joining the grounds together.
Many thanks
Andy
10-15-2012 06:02 AM
Hi Andy,
Do you wish to know this for the 4300 or 4330? Your title and the post both say different hardware types 🙂
Thank you,
10-15-2012 06:06 AM
It is the 4300 I am interested in - sorry about the typo.
Andy
10-15-2012
07:14 AM
- last edited on
09-22-2025
02:19 PM
by
Content Cleaner
Hi Andy,
You can see from the circuit board diagram on pg.6 of the spec sheet (found here) that each AI channel has its own COM port.
From the signal description table found in section 2-4 of the user manual (found here😞 "Analog Input Isolated Common Ground—These terminals are the reference point for differential analog input measurements."
You can therefore wire each AI port to a different ground. I hope this answers your question!
10-15-2012 08:18 AM
Yes that does.
Many thanks
Andy
10-15-2012 09:23 AM
You're welcome. I wish you all the best with your application!