PXI

cancel
Showing results for 
Search instead for 
Did you mean: 

PXIe-4331 Vch input impedance?

What is the input impedance of the Vch measurement amplifier?

I am interested in measuring a very high (100M-ohm+) resistance full bridge and I'm concerned a typical 1M-ohm input impedance will load the bridge but I can't find the specification in the documentation anywhere.

 

0 Kudos
Message 1 of 6
(3,230 Views)

From what I can see, NI does not specify the input resistance of the 4331.

 

What Voltage range are these measurements going to be in?

Chase
NI Technical Support Engineer
Message 2 of 6
(3,195 Views)

I prefer the lowest voltage, so that my digitizer resolution is as high as possible.  But one thing I saw which I thought was odd was the Vex "minimum current" at 29mA, which suggested my resistances are too high to work anyway...

0 Kudos
Message 3 of 6
(3,186 Views)

What resistance values are you using? Do you already have a 4331 on hand you are looking to use, or are you spec'ing a system?

Chase
NI Technical Support Engineer
Message 4 of 6
(3,172 Views)

I'm intending to use 1M-ohm resistances.  And I have ordered a system, but may need to return it if it is not suitable.

0 Kudos
Message 5 of 6
(3,163 Views)

I think this question would be better answered by a service request. Could you create one from your online account or call in so we can work a bit more efficently?

Chase
NI Technical Support Engineer
0 Kudos
Message 6 of 6
(3,146 Views)