I  consider the variant of the measurement temperature with help PXI-4071.

It is required the precision of measurement.

I found into the specification the RTD with accuracy 0.14 and Termistor with accuracy 0.08.

If I will use the termistor. What kind of termistor I have to use ?

If I will use RTD. Pt3851 RTD - is this type of resistance ? Where can I find that resistance ?

May be is the article about using the mode of temperature measument ?

Anybody used that method ?

 

Thanks for advance!

0 Kudos
Message 1 of 4
(2,622 Views)

Hi MaxiMuz,

 

According with the PXI-4071 specifications, you can use a thermistor in the range of -80 to 150 °C.

 

The pt stands for platinum and 3851 is the resistance at 0 °C in Ohms.

 

Regards,

Jos'e F.

0 Kudos
Message 2 of 4
(2,589 Views)

Do you mean, at 0°C its R=3851Ohm  ?

0 Kudos
Message 3 of 4
(2,583 Views)

Hi,

 

Yes, that is right.

 

Regards,

Jose