Calibrate NI 9265 module in cRIO
Hi guys
I have some NI 9265 analog current output modules, and I would like to perform an external calibration using a 8 1/2 digital multimeter available in the lab. I tried to do it using MAX interface, but I found out this is not possbile because the analog module is installed in a cRio 9074 controler. I only have this controler and an EtherCat chassi to install the modules. Is there some other way to perform the calibration using DAQmx with the NI 9265 installed in the cRio controler? Would it be possible in LabView?
(4,287 Views)
Accepted by topic author Lasn
Hey Lasn,
Unfortunately neither cRIO nor EtherCat chassis can be used with DAQmx  since they don't belong to DAQ platform. Some cSeries modules (e.g. NI 9265) are supported by both platforms, but it can be only calibrated by a cDAQ platform.
If you check the calibration manual for the NI 9265 (if you hadn't checked yet) you'll see that it recommends you to use a cDAQ-9178, which a 8 slot cDAQ chassis.
http://www.ni.com/pdf/manuals/372667b.pdf
If you really need an external calibration, please contact your local NI office to get more information on Calibration Services.
Best Regards.
Felipe Flores
Technical Support Engineer
(4,254 Views)
Thanks, Felipe_vInst. I was afraid of that. I will try to get my hands on a cDaq chassi.
(4,245 Views)
Ok Lasn,
Let us know in case you need help with the Calibration process when you get the cDAQ chassis.
Best Regards
Felipe Flores
(4,222 Views)